This is the report for 4/29/19. All stream/station now list the baseline turbidity for that site, to help you identify whether turbidity elevations during baseflow indicate the need for cleaning. The baseline turbidity NTU appears in the header of the history of each site that follows the tables. Baseline was reasonably estimated by examining turbidity patterns after installation.
@lawler SL101(Musconetcong U) Turbidity has risen to 1300-1400 NTU and sensors may be packed or embedded. This site fouls easily, but the extent of fouling could possibly cause damage.
@bilger SL117(SandyRun) The battery is in a pattern of decline (~3.6 v) which is below recommended level for logger stability.
@bilger SL118(Wissahickon) The battery here also needs to be checked as it is 3.1-3.2v, and recharged. Check for moisture on the battery and inside the Pelican box.
@madden SL122(Buckwha D) Sensors were naturally cleaned but need more cleaning as turbidity is well over 10 x the natural level at baseline flow. Sensors do not appear to have been cleaned this winter and spring.
@ebel SL131(Little Lehigh) Any update on this site?
@weinberg SL134(Mine Hole) There is major fouling of the sensors since april 21. This station fouls easily and needs a volunteer who can check it with some regularity.